[Cytometry] multiple negatives for comp in Diva?

Ian Titley Ian.Titley at icr.ac.uk
Fri Feb 7 09:03:16 EST 2014


Yep. However if Carol is using cells for the GFP channel and beads for say the PE channel will the fact that the autoflourescence of the two types of particle does not match in both channels impact the compensation between those channels?

Ian

-----Original Message-----
From: Cris Bare [mailto:flowmail at verizon.net] 
Sent: 07 February 2014 13:52
To: Ian Titley
Cc: Carol Norris; cytometry at lists.purdue.edu
Subject: Re: [Cytometry] multiple negatives for comp in Diva?

Let me see of I can beat Mario this time. 

The autofluorescence ( AKA intrinsic fluorescence) should match between the positive population and the negative population within the same detector channel. 

For example, if you capture PECy7 on a bead, use the corresponding bead negative, but if you use GFP the negative should be the same cell type non-expressor. 

That's why I suggest manually drawing the "P3" negatives gate in the histogram instead of using the universal unstained negative. 

-cbb

On Feb 7, 2014, at 1:29 AM, Ian Titley <Ian.Titley at icr.ac.uk> wrote:

> Will this work if the "autofluoresence" intensity of the "negative" beads and "negative" cells are significantly different within the same compensation experiment? I think this might be what Carol is concerned about? Can Carol use antibody labelled beads as "positives" and unlabelled cells as the universal negative control within the same compensation run?
> 
> Best wishes
> 
> Ian
> 
> 
> Ian Titley PhD
> The Institute of Cancer Research | Centre for Evolution and Cancer | 
> Division of Molecular Pathology | Brookes Lawley Building | 15 
> Cotswold Road | Sutton | Greater London | SM2 5NG
> 
> Tel: +44 (0)20 8722 4255 | Email:  ian.titley at icr.ac.uk| Web:  
> www.icr.ac.uk | Twitter @ICRnews For regular updates on the ICR, 
> please follow us on Facebook: 
> www.facebook.com/theinstituteofcancerresearch
> 
> Making the discoveries that defeat cancer
> 
> 
> 
> 
> 
> 
> -----Original Message-----
> From: cytometry-bounces at lists.purdue.edu 
> [mailto:cytometry-bounces at lists.purdue.edu] On Behalf Of Cris Bare
> Sent: 06 February 2014 17:41
> To: Carol Norris
> Cc: cytometry at lists.purdue.edu
> Subject: Re: [Cytometry] multiple negatives for comp in Diva?
> 
> Carol,
> 
> It's easier than you think. 
> 
> In DiVa on the comp controls worksheet in the fluorochrome specific histogram where you adjust gate P2 to select the positive population, you can draw a P3 gate on the negative population and DiVa will use P3 preferentially for calculation. 
> 
> That's a word sentence. 
> 
> 1. Include negatives in each comp control tube. 
> 2. Add a P3 gate on the negatives in the same histogram as P2 positives. 
> 3. Calculate. 
> 
> This is why DiVa has the checkbox when you create your comp tubes to include or not include a universal negative. 
> 
> Cheers!
> -cbb
> 
> On Feb 6, 2014, at 8:40 AM, Carol Norris <carol.norris at uconn.edu> wrote:
> 
>> Hi Flow Gurus,
>> 
>> We have lots of folks combining GFP-expressing cells with antibody labeling, and I’d like to get more of them using comp beads for the antibody controls. There doesn’t seem to be a way to use GFP-negative cells for the GFP comp control and unstained beads for the antibody comp controls in Diva.  Are unlabeled cells OK to use as a universal negative?  The other option would be to transfer the data to FlowJo to get a comp matrix, and then transfer the comp values back into Diva.
>> 
>> Thanks,
>> Carol
>> 
>> 
>> Carol E. Norris, Ph.D
>> Facility Scientist
>> Flow Cytometry/Confocal Microscopy Facility Biotechnology/Bioservices 
>> Center University of Connecticut Unit 3149
>> 91 N. Eagleville Rd
>> Storrs, CT 06269-3149
>> 
>> Phone (860) 486-3080
>> Fax (860) 486-5005
>> 
>> _______________________________________________
>> Cytometry mailing list
>> Cytometry at lists.purdue.edu
>> https://lists.purdue.edu/mailman/listinfo/cytometry
>> Search the list archive at  http://tinyurl.com/cytometry
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu
> https://lists.purdue.edu/mailman/listinfo/cytometry
> Search the list archive at  http://tinyurl.com/cytometry
> 
> The Institute of Cancer Research: Royal Cancer Hospital, a charitable Company Limited by Guarantee, Registered in England under Company No. 534147 with its Registered Office at 123 Old Brompton Road, London SW7 3RP.
> 
> This e-mail message is confidential and for use by the addressee only.  If the message is received by anyone other than the addressee, please return the message to the sender by replying to it and then delete the message from your computer and network.

The Institute of Cancer Research: Royal Cancer Hospital, a charitable Company Limited by Guarantee, Registered in England under Company No. 534147 with its Registered Office at 123 Old Brompton Road, London SW7 3RP.

This e-mail message is confidential and for use by the addressee only.  If the message is received by anyone other than the addressee, please return the message to the sender by replying to it and then delete the message from your computer and network.



More information about the Cytometry mailing list