"Back to the original posting, it seems clear from responses and brief
literature search that 5&10 double positive cases exist and shouldn't be
discounted as "incorrect" antibody. Brian's issue with one clone positive
and one clone negative points to a technical not biological issue with that
vial of that clone in that lab during that test. Hopefully he resolves it
and can control for it in future panels."

The problem is just one: to identify mature B cell from one B lymphoma
CD10+ you can not use all the clones available in the market. The best
options are ALB1 and RD1-J5 from Coulter. Anyway with all clones available
you can identify neutrophil CD10+ and immature B cells, which means you can
use all the others to identify B-ALL.

Hope that helps.

Nuno Oliveira, IPOCFG Portugal - Flow Cytometry division


2014-02-01 Nuno Oliveira <cravista at gmail.com>:

> "Back to the original posting, it seems clear from responses and brief
> literature search that 5&10 double positive cases exist and shouldn't be
> discounted as "incorrect" antibody. Brian's issue with one clone positive
> and one clone negative points to a technical not biological issue with that
> vial of that clone in that lab during that test. Hopefully he resolves it
> and can control for it in future panels."
>
> The problem is just one: to identify mature B cell from one B lymphoma
> CD10+ you can not use all the clones available in the market. The best
> options are ALB1 and RD1-J5 from Coulter. Anyway with all clones available
> you can identify neutrophil CD10+ and immature B cells, which means you can
> use all the others to identify B-ALL.
>
> Hope that helps.
>
> Nuno Oliveira, IPOCFG Portugal - Flow Cytometry division
>
>
>
> 2014-02-01 Cris Bare <flowmail at verizon.net>:
>
> Thanks for the clarification, Nuno. More markers = better definition!
>>
>> Back to the original posting, it seems clear from responses and brief
>> literature search that 5&10 double positive cases exist and shouldn't be
>> discounted as "incorrect" antibody. Brian's issue with one clone positive
>> and one clone negative points to a technical not biological issue with that
>> vial of that clone in that lab during that test. Hopefully he resolves it
>> and can control for it in future panels.
>>
>> -cbb
>>
>> On Feb 1, 2014, at 2:26 AM, Nuno Oliveira <cravista at gmail.com> wrote:
>>
>> > Hi
>> >
>> > Not all hairy cell leukemias have CD5+ and CD10+. Normally both should
>> be
>> > negative. In some cases, one or both could appear positive. You can also
>> > use LAIR1 and CD200 to help on this. Both very strong means hairy cell
>> > leukemia with CD20 strong among other markers...
>> >
>> > Yours sincerely,
>> >
>> > Nuno Oliveira, IPOCFG Portugal - Flow Cytometry division
>> >
>> >
>> >
>> > 2014-01-31 Cris Bare <flowmail at verizon.net>:
>> >
>> >> From years ago I have this notion that CD5 CD10 double positive is
>> hairy
>> >> cell lymphoma. The name stuck with me.
>> >>
>> >> In the panel are you running single color anti-CD10 tubes or
>> multicolor?
>> >> Could there be a setup issue where one or the other wasn't properly
>> gated
>> >> or controlled?
>> >>
>> >> Personally I'd agree with you that stopping the use of one clone over
>> one
>> >> case is a bit rushed. But I'm also not an MD.
>> >>
>> >> -cbb
>> >>
>> >>
>> >> On Thu, Jan 30, 2014 at 7:58 AM, Marsiglia, Brian <
>> >> Brian.Marsiglia at inova.org
>> >>> wrote:
>> >>
>> >>> I had an interesting case this week.  It was a tissue that was CD5,
>> CD10
>> >>> positive (B cell Lymphoma). I went back and reset the CD10 tubes and
>> got
>> >>> the same result. In my panel I have two CD10 clones, which are also
>> >>> different fluorochromes. The CD10 in FITC (ALB 2 clone) was CD10
>> >> negative,
>> >>> and the CD10 PC5 (ALB 1 clone) was CD10 positive.
>> >>>
>> >>> The clinician involved in this case has ordered immunostains to verify
>> >>> which is "correct" .Clinically this case goes with a CD5 positive
>> >>> immunophentype.  It was also requested that I stop using whichever
>> clone
>> >> is
>> >>> the "incorrect" clone, confirmed my immunostains.
>> >>>
>> >>> I'm not a fan of this idea.
>> >>>
>> >>> I've done a good deal of reading on this topic, and it looks like
>> there
>> >>> are cases where the lymphomas can be positive for both.
>> >>>
>> >>>
>> >>> Any feedback on this situation would be appreciated.
>> >>>
>> >>> Thanks!
>> >>> Brian
>> >>>
>> >>> _______________________________________________
>> >>> Cytometry mailing list
>> >>> Cytometry at lists.purdue.edu
>> >>> https://lists.purdue.edu/mailman/listinfo/cytometry
>> >>> Search the list archive at  http://tinyurl.com/cytometry
>> >>>
>> >>
>> >> _______________________________________________
>> >> Cytometry mailing list
>> >> Cytometry at lists.purdue.edu
>> >> https://lists.purdue.edu/mailman/listinfo/cytometry
>> >> Search the list archive at  http://tinyurl.com/cytometry
>> >>
>> >>
>> > _______________________________________________
>> > Cytometry mailing list
>> > Cytometry at lists.purdue.edu
>> > https://lists.purdue.edu/mailman/listinfo/cytometry
>> > Search the list archive at  http://tinyurl.com/cytometry
>>
>
>


More information about the Cytometry mailing list