[Cytometry] Colour compensation

Vivek Tanavde vtanavde at hotmail.com
Thu Apr 18 06:11:13 EDT 2013


Hello Rabab,Why would you want to do compensation on side scatter? As far as I know compensation is necessary for bleed through of a fluorescent dye in other PMTs detecting fluorescence. Since the SSC detector does not detect fluorescence, it is not necessary to compensate the signal on this detector.Regards,


Vivek Tanavde 

Bioinformatics Institute
Singapore

From: rababhaider3 at hotmail.com
To: cytometry at lists.purdue.edu
Date: Tue, 16 Apr 2013 19:18:33 +0400
Subject: [Cytometry] Colour compensation

Hi everyone! 
I'm using FC500. 
Is there anyone uses FC500 and knows how to do the automated compensation on SS Log scale. All the options from auto setup schedule in the software uses FS Lin/ SS Lin. 
Thank you, 
Regards, 
Rabab Al lawati 
Royal Hospital 
Oman 
 		 	   		  