[Cytometry] LSRII sample pressure issue

prateek arora exquisiteprateek at rediffmail.com
Fri May 20 04:18:21 EDT 2011


Hi lin,

The back flow of sheath fluid in to your sample tube could be because 
of following reasons-
- Either the FACS tube would be crack somewhere so that sample pressure 
is leaking from the tube or the FACS tube using by you might be not 
compatible thus unable to pressurize the tube.
- Wrong connection of waste tubing due to connecting the waste tubing 
in to the air vent port rather than in to the waste vent port of the 
waste reservoir. (the waste tube should be connected to the waste vent 
port of the waste reservoir.)
- During shut downing of your instrument, use of excess of water that 
can cause revert back flow of liquid in to the sample. Around 1ml of 
water should be remaining in the the FACS tube not more than that.
- Either the Pressure pump is not functioning properly.
- If waste reservoir is not properly closed.

Try to look at this following major points of causing of back flow. I 
hope it would work and help you to troubleshoot out with this problem.

Good luck.

Prateek Arora




On Fri, 20 May 2011 12:13:32 +0530  wrote
>Hi Lin,



The phenomenon that sheath is leaking back in your sample sounds

familiar, although in our case it happened directly, not after 30min.

We've had this issue a couple of times, when users placed tubes with to

much liquid. When pressure is put on the tube, this causes liquid to

creep up in the tubing for the sample pressure and wet the filter

inside. The strange thing here is that you would have the leaking back

immediately if it were wet filters, not after 30min. If you place a

tube with FACSflow with the table not under the tube, does the level

drop? 

Maybe the tubing to the waste is somewhat clogged, so after 30min of

analysis the resistance to pump waste becomes to high? This would be my

best guess, hope it helps.



Good luck,

Floris





Floris van Alphen

Dept. of Molecular Cell Biology/

Central facility for cell analysis

Sanquin Bloodsupply, research division

room U264A

Plesmanlaan 125, 1066CX Amsterdam

the Netherlands

+31(0)20-5123250

F.vanalphen at sanquin.nl







-----Original Message-----

From: cytometry-bounces at lists.purdue.edu

[mailto:cytometry-bounces at lists.purdue.edu] On Behalf Of

lzhang11 at ncsu.edu

Sent: donderdag 19 mei 2011 23:05

To: Cytometry at lists.purdue.edu

Subject: [Cytometry] LSRII sample pressure issue





Hey flower,



We have low sample pressure problem (running fine for 30 min and then

drop to 0-20 event/sec)and we observe fluids filled in to the sample

tube. This is still happening even after PM and having DCM tubing

changed and as much as other valves and filters changed.



We do not have contract service so it is expensive to replace the whole

pneumatic and fluidics system. Can I ask you guys if you have similar

problem but get fixed with replacement with any particular part? Our

LSRII is 6 year old.



Thanks,



lin





_______________________________________________

Cytometry mailing list

Cytometry at lists.purdue.edu

https://lists.purdue.edu/mailman/listinfo/cytometry

Search the list archive at http://tinyurl.com/cytometry







Disclaimer:

http://www.sanquin.nl/Sanquin-nl/sqn_disclaimer_nl.nsf/All/Disclaimer-
E-Mail-Sanquin.html