Thanks Joe - I guess we would assess linearity in practice, preferentially by using software such as CS&T, or manually using beads and ploting PMT values versus say MFI for each detector?

Ian


Ian Titley PhD
Section of Haemato-oncology
Institute of Cancer Research
15 Cotswold Road
Belmont
Sutton
Surrey  SM2 5NG
UK

Tel +44 (0)20 8722 4255



>>> <Joe_Trotter at bd.com> 15/03/2011 17:08 >>>
Ian,

        I think we should refine that, and say that since linearity does 
matter when sampling bright single color controls (as input to a spill 
matrix for compensation), that the positives can be as bright as needed as 
long as they are on-scale within the linear range of the detector.

        Joe


Joseph Trotter
Principal Scientist, BD Fellow / Advanced Technology Group

BD Biosciences
tel: (858) 812-8946  cell: (858) 449-7338  fax: (858) 812-8968
10975 Torreyana Road, San Diego, CA 92121-1106 USA 
E-mail: Joe_Trotter at bd.com   Website: www.bd.com 




cytometry-bounces at lists.purdue.edu wrote on 03/15/2011 08:21:55 AM:

> From: "Ian Titley" <Ian.Titley at icr.ac.uk>
> To: "Mario Roederer" <roederer at drmr.com>, "Helen Ferry" 
> <helen.ferry at ndm.ox.ac.uk>, "Ian Dimmick" 
> <ian.dimmick at newcastle.ac.uk>, "Mary A' 'Price" <mprice1 at tulane.edu>
> Cc: "cytometry at lists.purdue.edu" <cytometry at lists.purdue.edu>
> Date: 03/15/2011 08:27 AM
> Subject: Re: [Cytometry] compensation
> Sent by: cytometry-bounces at lists.purdue.edu 
> 
> Absolutely, in fact brighter is better (more photons to play with) 
> as long as it stays on scale!
> 
> Ian
> 
> Ian Titley PhD
> Section of Haemato-oncology
> Institute of Cancer Research
> 15 Cotswold Road
> Belmont
> Sutton
> Surrey  SM2 5NG
> UK
> 
> Tel +44 (0)20 8722 4255
> 
> 
> 
> >>> Ian Dimmick <ian.dimmick at newcastle.ac.uk> 15/03/2011 14:59 >>>
> Incorrect , compensation is not affected by signal intensity , so 
> you are safe 
> 
> 
> Ian
> 
> -----Original Message-----
> From: cytometry-bounces at lists.purdue.edu [mailto:cytometry-
> bounces at lists.purdue.edu] On Behalf Of Price, Mary A
> Sent: 15 March 2011 14:36
> To: Helen Ferry; Mario Roederer
> Cc: cytometry at lists.purdue.edu 
> Subject: Re: [Cytometry] compensation
> 
> Hi Everyone,
> 
> On this thread, I have a question that might be considered basic but
> haven't seen it asked or answered (since it might be very basic and I
> should know it) about the beads and staining. I have yet to use the
> beads or have others to do it since I am concerned that since beads are
> uniform, they will bind the antibody better then a cell will since it
> has more receptors for it to bind to. My concern is that it will
> actually show a brighter signal then what a cell with less binding would
> show therefore throwing off the compensation. Is this not a correct
> assumption? 
> 
> Mary
> 
> -----Original Message-----
> From: cytometry-bounces at lists.purdue.edu 
> [mailto:cytometry-bounces at lists.purdue.edu] On Behalf Of Helen Ferry
> Sent: Tuesday, March 15, 2011 5:02 AM
> To: Mario Roederer
> Cc: cytometry at lists.purdue.edu 
> Subject: Re: [Cytometry] compensation
> 
> Hi 
> 
> Diva does allow you to have a negative population for each compensation
> control. You just need to create a new region for the negative
> population in each of the single-stained controls and delete the
> unstained tube from the compensation matrix. This enables you to have a
> mixture of cells and beads in your compensation controls panel although,
> obviously, you should use either beads or cells for the positive and
> negative in a single control.
> 
> Best wishes
> Helen
> 
> -----Original Message-----
> From: cytometry-bounces at lists.purdue.edu 
> [mailto:cytometry-bounces at lists.purdue.edu] On Behalf Of Mario Roederer
> Sent: 14 March 2011 23:52
> To: cytometry at lists.purdue.edu 
> Subject: Re: [Cytometry] compensation
> 
> Again, I'd like to stress that there are only 3 rules to achieve proper
> compensation:
> 
> For each color in any given panel:
> 
> (1) Your compensation control color must be matched to your experimental
> color
> (2) The positive control must be at least as bright as your experimental
> sample
> (3) The positive control for your compensation sample, when unstained,
> must have the same background fluorescence as the negative control for
> that color.
> 
> That's it. 
> 
> You'll note that there is nothing about PMT voltages, filters,
> autofluorescence, cell type, beads, day of the week, phase of the moon,
> or mood of the operator involved.
> 
> All reasonable problems with compensation can be traced to a violation
> of one of these 3 rules.
> 
> I will note that (current) DiVa software enables a common violation of
> rule #3, by only allowing only one universal negative -- thus preventing
> proper compensation when a mixture of bead-based and cell-based
> compensation controls are used.  Better-designed software does not make
> this fundamental mistake.
> 
> mr
> 
> (PS, I left out one rule: that you use the same instrument fluorescence
> settings for all samples and compensation controls.  Violating this rule
> is more fundamental than anything compensation-related; including it as
> a rule would be akin to specifying that you must turn on the instrument
> before you run any controls.  This would be what I would refer to as an
> "unreasonable problem" with compensation.)
> 
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu 
> https://lists.purdue.edu/mailman/listinfo/cytometry 
> Search the list archive at  http://tinyurl.com/cytometry 
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu 
> https://lists.purdue.edu/mailman/listinfo/cytometry 
> Search the list archive at  http://tinyurl.com/cytometry 
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu 
> https://lists.purdue.edu/mailman/listinfo/cytometry 
> Search the list archive at  http://tinyurl.com/cytometry 
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu 
> https://lists.purdue.edu/mailman/listinfo/cytometry 
> Search the list archive at  http://tinyurl.com/cytometry 
> 
> The Institute of Cancer Research: Royal Cancer Hospital, a 
> charitable Company Limited by Guarantee, Registered in England under
> Company No. 534147 with its Registered Office at 123 Old Brompton 
> Road, London SW7 3RP.
> 
> This e-mail message is confidential and for use by the addressee 
> only.  If the message is received by anyone other than the 
> addressee, please return the message to the sender by replying to it
> and then delete the message from your computer and network.
> 
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu 
> https://lists.purdue.edu/mailman/listinfo/cytometry 
> Search the list archive at  http://tinyurl.com/cytometry 


-----------------------------------------
*******************************************************************
IMPORTANT MESSAGE FOR RECIPIENTS IN THE
U.S.A.:
This message may constitute an advertisement of
a BD group's products or services or a
solicitation of interest in them. If this is
such a message and you would like to opt out of
receiving future advertisements or
solicitations from this BD group, please
forward this e-mail to optoutbygroup at bd.com.
*******************************************************************
This message (which includes any attachments)
is intended only for the designated
recipient(s).  It may contain confidential or
proprietary information and may be subject to
the attorney-client
privilege or other confidentiality protections.
 If you are not a designated recipient, you may
not review, use, copy or distribute this
message. If you received this in error, please
notify the sender by reply e-mail and delete
this message. Thank you.
*******************************************************************
Corporate Headquarters Mailing Address: BD
(Becton, Dickinson and Company) 1 Becton Drive
Franklin Lakes, NJ 07417 U.S.A.
*******************************************************************

The Institute of Cancer Research: Royal Cancer Hospital, a charitable Company Limited by Guarantee, Registered in England under Company No. 534147 with its Registered Office at 123 Old Brompton Road, London SW7 3RP.

This e-mail message is confidential and for use by the addressee only.  If the message is received by anyone other than the addressee, please return the message to the sender by replying to it and then delete the message from your computer and network.



More information about the Cytometry mailing list