Hi Jeff,

Being able to identify and sort a derivative chromosome will depend on it
having a different size (DNA content) or bp ratio such that it would move
from its normal position on a flow karyotype. Small changes, reciprocal
translocations and so on are often not seen. Traditionally (and we haven't
sorted a chromosome in anger for quite a while) we used H33258 and
chromomycin A3. As you say chromomycin is best excited using a 457nm line (a
rather attractive blue) and I recall using around 300mW of that and UV to
get best resolution, the more you put in with small stuff the better the
signal and CV. Havent tried with a violet laser but I suspect the power
output wouldn't give you the required resolution. I think there have been
papers using 488nm excited dyes but again I don't think they work that well
and the Hoechst/chromomycin combination is still the gold standard.

Best wishes,
Derek



On 16/7/09 08:55, "Jeff Barry" <JBarry at picr.man.ac.uk> wrote:

> Hello,
> 
>  
> 
> We have a researcher who would like to resolve and sort human
> isochromosome 17q.  From the literature I found that a bivariate plot of
> Hoechst against Chromomycin A3 seems to resolve chromosome 17 nicely but
> does anyone know if the i(17q) will form a distinct separate population?
> 
>  
> 
> I'm also concerned that our present sorters' configurations may not be
> ideally suited to this, especially regarding the choice of laser lines
> and power. We have an Influx with 100mw UV and 50wm 407nm laser. I
> understand that the laser line of choice for Chromomycin is 457nm but
> has anyone tried with a 407nm line?
> 
>  
> 
> We also have a Vantage that has a UV laser (<100mw) and an Argon ion
> laser that could in theory be tuned to 457nm (100wm).
> 
>  
> 
> Does anyone have an opinion whether these sorters could be used?
> 
>  
> 
> If anyone has done this before and is willing to share their experience
> I would be very grateful.
> 
>  
> 
> Jeff Barry
> 
> Cancer Research UK
> 
> Paterson Institute
> 
> Manchester University
> 
> England 
> --------------------------------------------------------
> This email is confidential and intended solely for the use of the person(s)
> ('the intended recipient') to whom it was addressed. Any views or opinions
> presented are solely those of the author and do not necessarily represent
> those of the Paterson Institute for Cancer Research or the University of
> Manchester. It may contain information that is privileged & confidential
> within the meaning of applicable law. Accordingly any dissemination,
> distribution, copying, or other use of this message, or any of its contents,
> by any person other than the intended recipient may constitute a breach of
> civil or criminal law and is strictly prohibited. If you are NOT the intended
> recipient please contact the sender and dispose of this e-mail as soon as
> possible.  
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu
> https://lists.purdue.edu/mailman/listinfo/cytometry

-- 
***************************************************************
Derek Davies, FACS Laboratory, London Research Institute,
Cancer Research UK, 44 Lincolns Inn Fields, London, UK.

Tel: (44) 20 7269 3394
FAX: (44) 20 7269 3479
Mobile: 07790 604112
e_mail: derek.davies at cancer.org.uk
http://science.cancerresearchuk.org/sci/facs/

In tenebris lux    
***************************************************************


This communication is from Cancer Research UK. Our website is at www.cancerresearchuk.org. We are a charity registered under number 1089464 and a company limited by guarantee registered in England & Wales under number 4325234. Our registered address is 61 Lincoln's Inn Fields, London WC2A 3PX. Our central telephone number is 020 7242 0200.
 
This communication and any attachments contain information which is confidential and may also be privileged.   It is for the exclusive use of the intended recipient(s).  If you are not the intended recipient(s) please note that any form of disclosure, distribution, copying or use of this communication or the information in it or in any attachments is strictly prohibited and may be unlawful.  If you have received this communication in error, please notify the sender and delete the email and destroy any copies of it.
 
E-mail communications cannot be guaranteed to be secure or error free, as information could be intercepted, corrupted, amended, lost, destroyed, arrive late or incomplete, or contain viruses.  We do not accept liability for any such matters or their consequences.  Anyone who communicates with us by e-mail is taken to accept the risks in doing so.


More information about the Cytometry mailing list