[Cytometry] Compensation with liver cells

Mario Roederer roederer at drmr.com
Mon Jul 13 08:31:13 EDT 2009


For those of you who are math-averse, or compensation-averse, you can  
skip this entire email!  However, the only way to put this issue to  
rest will be to go in deteail through the compensation math.

Simon, you are correct that, on older analog instruments, when the  
negative bead population is on the lower axis, the median will be  
reported as 1, as  you note.  And, this median value is incorrectly  
high.  Presumably, it should be something like.... 0.3 or 0.1.  Let's  
assume, in the most extreme case, that the value should have been 0.  
(Since it can't be negative, this is as much inaccuracy as is possible).

Let's further assume that you have turned down the voltages to get  
your cells on-scale (i.e., between 10^3 and 10^4), and that you have  
correctly used beads whose positive fluorescence is at least as bright  
as your cells.  Therefore, your positive bead median fluorescence is  
(at least) 10^3 (1000).  We'll use 1000 as the "worst case"  
possibility.  (This is the worst case, because if it is lower than  
this, you need to turn up the PMT voltage and then hopefully your  
negatives will come up off the axis as well).

The compensation coefficient is the ratio of the DIFFERENCE of the  
MFI's (of the spillover channel and the primary channel).  Thus, in  
this channel, the difference will be computed as "positive - negative  
MFI" = "1000 - 1", which, as you note, is slightly incorrect (it  
should be, in our extreme case, "1000 - 0").  Assume that the  
difference in MFI of the positive and negative bead population in the  
spillover channel is "dS" (delta-spillover).

Thus, the compensation coefficient will be computed as

ds / (1000 - 1) = dS / 999.

instead of the exact

ds / (1000 - 0) = dS / 1000.

You will note that the difference in these values is one part in  
1000.  Thus, if the correct spillover value was supposed to be 15.0%  
(i.e., dS = 150), by having the beads on the axis, it now comes out to  
15.015%.

There is no meaningful difference between these two compensation  
values.  In fact, the precision on the compensation measurement is far  
less than this difference -- i.e., the typical variance on the  
computation of a compensation coefficient will be on the order of  
several percent, and, unless you are collecting tens of thousands of  
beads for each compensation control, may well be 5-10% or more.  (This  
error is primarily driven by the imprecision of the median of the  
positive, bright beads).  Since this is one to two orders of magnitude  
greater than the inaccuracy introduced by pegging the negative beads  
on the axis, the contribution of the latter is negligible.

The bottom line is that because the compensation coefficients are  
computed based on the DIFFERENCE in MFI's (and not the RATIO of  
MFI's), small absolute errors in the position of the negative control  
become irrelevant if the positive controls are nice and bright.  The  
error in the compensation coefficient is the sum of the absolute  
errors in the MFI's of both the negative and the positive control; the  
latter has an inherently MUCH larger absolute error than the former.

mr


(Incidentally, if you peg the negative beads in the spillover channel  
as well, and there is very low spillover (dS is <10), then the error  
in the compensation coefficient might be much larger.  However, as  
long as your compensation control is at least as bright as your cells  
(this should be your mantra), then even this larger error will not  
affect your measurement, because the absolute magnitude of this error  
on the cells will be much less than the variance in the  
autofluorescence of the cells.)




On Jul 13, 2009, at 5:55 AM, SIMON MONARD wrote:

> Ian
> All the instruments in my lab are NOT digital. Probably most
> instruments in the world are not digital. Granted you could use comp
> beads on digital machines when looking at large cells. However, on a
> Calibur, as soon as 50% of the blank beads are in the first channel
> the median will be 1. Continue to reduce the voltages and it will
> remain 1. Likewise your positive beads will have a median of 1 when
> 50% are in the first channel and will continue to be 1 as you increase
> the compensation. So, the compensation will apparently be correct for
> a wide range of PMT voltages and compensation settings. which ever way
> you visualize and calculate compensation this will not change the fact
> you have all those uninformative ones.
> I am suggesting an inexpensive solution that works with all platforms
> that can be used with real time or software compensation.
>
> Simon
>
>  Quoting Ian Dimmick <ian.dimmick at newcastle.ac.uk>:
>
>> Beads can not go off scale , the relative negative events will
>> allways be visualisable to whatever degree, it has been my
>> experience that you can always compensate on the calibur using these
>> beads , you would stop your compensation when the median of the
>> bright was the same as the negative , or if you wish you could
>> import the Calibur data into diva and use the bi exponential scale
>> to visualise a little better the negative compensation beads ,
>> although still using the median.I did not ignore Simon's comment
>> regarding the autofluorescence , if you re read my mail I stated
>> that autofluorescence is taken into account in the inverted matrix
>> calculation , whatever value it has , you defend Simon but I do not
>> feel his position needs to be defended , we all have our ways of
>> doing things and we all have our views on "the best" way to do
>> things, this is why this forum thrives, in many scenarios we are all
>> helped by people offering their comments on personal experience ,
>> and very ofte!
>> n those views will differ , primarily because flow cytometry ,cells
>> , filters , lasers....................are not perfect , and very
>> far  from standard, sometimes even within the same instrument
>> platform ,  so please if you want to make comments then make them in
>> context of  your own experience , do not miss quote, I felt that the
>> whole  discussion was starting to have the effect of confusing
>> rather than  helping , I did not aim that at Simon , but the whole
>> thread in  general , however if Simon felt that my comment was aimed
>> at his  mail the I appolgise to him .
>> I do however stand by my comments
>>
>>
>> Ian Dimmick
>> Flow Cytometry Core Facility Manager
>> Institute of Human Genetics
>> Bioscience Centre
>> International Centre for life
>> Newcastle Upon Tyne
>> NE1 3BZ
>> UK
>> Ian.Dimmick at ncl.ac.uk
>> Tel 0044 191 2418831
>> Fax 0044 191 2418666
>> (mob) 0044 7970344823
>> http://www.ncl.ac.uk/ihg
>> ________________________________________
>> From: cytometry-bounces at lists.purdue.edu
>> [cytometry-bounces at lists.purdue.edu] On Behalf Of Ray Hicks
>> [rhicks at cytekdev.com]
>> Sent: 11 July 2009 16:07
>> To: 'Mario Roederer'; cytometry at lists.purdue.edu
>> Subject: Re: [Cytometry] Compensation with liver cells
>>
>> So Mario and Ian,
>>
>> On an analogue machine, say a Calibur, if you wish to set  
>> compensation for
>> bright cells, you'd set your PMT voltages so that the negative  
>> cells would
>> be on scale, then you pop on your compensation beads, the  
>> unlabelled comp
>> beads are off scale or clipped to the origin of the graph, the  
>> stained comp
>> beads are on scale, you start compensating them so that the spillover
>> fluorescence is equivalent to that on the unstained bead, you hit a  
>> pitfall,
>> in that you don't know when to stop.  Whether you use mean or  
>> median or
>> magic, you're not going to know when you've applied enough  
>> compensation,
>> since you don't know where the negatives lie.  Using compensation  
>> beads that
>> have enough "Autofluorescence" to be on scale at the same time as the
>> stained ones is pretty much essential I'd say under this  
>> circumstance - one
>> that Simon described and which you've both ignored in your answers.
>>
>> Care would probably have to be taken that the bead autofluorescence  
>> was
>> neutral over the wavelengths investigated.
>>
>> I don't think Simon was confusing any _issues_ Ian
>>
>> Ray
>>
>> Ray Hicks
>>
>> http://www.cytekdev.com
>>
>> Tel: +44 (0)208 1337 968  Fax: +44(0)208 5889 004  Skype:  
>> ray.hicks.cytek
>>
>> Cytek Development Europe, Unit 8 The Maltings, Millfield, Cottenham,
>> Cambridge CB24 8RE. UK
>>
>>
>>
>> -----Original Message-----
>> From: cytometry-bounces at lists.purdue.edu
>> [mailto:cytometry-bounces at lists.purdue.edu] On Behalf Of Mario  
>> Roederer
>> Sent: 10 July 2009 20:43
>> To: cytometry at lists.purdue.edu
>> Subject: Re: [Cytometry] Compensation with liver cells
>>
>> As I wrote, as long as your positively-stained beads are at least as
>> bright as your stained cells, they are a perfectly good compensation
>> control.  In today's instrumentation, it's irrelevant if the negative
>> (unstained) beads are "not onscale".  With modern digital  
>> electronics,
>> it's not even possible, since all signals are collected in linear,  
>> for
>> which even zero measured fluorescence is still onscale.
>>
>> So in fact, the background fluorescence of compensation beads is not
>> relevant, and there is no need to match it to the cells. Don't worry
>> if the beads are pegged at the bottom when you run on the voltages
>> optimized for your cells.
>>
>> So.... In summary:  No!
>>
>> mr
>>
>> On Jul 10, 2009, at 12:48 PM, SIMON MONARD wrote:
>>
>>> No...?
>>>
>>> If you are looking at larger cells, ESCs, Macs, hepatocytes etc you
>>> will find that setting your voltages on the negative cells the beads
>>> will be offscreen. Bump up the voltages enough to get the beads on
>>> screen and you may find you have little dynamic range left to  
>>> measure
>>> fluorescence. You can find that your negative cells are two logs
>>> higher than blank comp beads. If you have an older machine with 4  
>>> log
>>> decades BD comp beads are useless if you work with such cells. The  
>>> BD
>>> beads have clearly been selected because they have a similar
>>> background to lymphocytes. Larger beads have a higher background but
>>> also capture more antibody so they can be a lot brighter than the
>>> commercial beads. I'm not suggesting adding background  
>>> fluorescence to
>>> blank beads I'm suggesting selecting beads that better match your
>>> cells and coupling them to anti-rat Ig or whatever. 6 micron
>>> polycarbonate beads are a much better choice for larger cells.
>>> In summary...Yes!
>>>
>>> Simon
>>>
>>>
>>> Quoting Mario Roederer <roederer at drmr.com>:
>>>
>>>> No... antibody capture beads are not designed with a particular  
>>>> cell
>>>> subset in mind.  Any compensation bead will work for any cell type,
>>>> as
>>>> long as the positive beads are brighter (or at least as bright) as
>>>> your stained cells.
>>>>
>>>> Adding background to your beads (or using beads with more  
>>>> background)
>>>> will only decrease the precision (although not the accuracy) of  
>>>> your
>>>> compensation.
>>>>
>>>> mr
>>>>
>>>>
>>>> On Jul 10, 2009, at 5:24 AM, SIMON MONARD wrote:
>>>>
>>>>> You will likely find them unusable as the commercial ones are
>>>>> designed
>>>>> for use with lymphocytes. If you are planning on doing much FACS
>>>>> then
>>>>> I suggest you make your own capture beads using beads with a
>>>>> background fluorescence comparable to your unstained cells
>>>>
>>>> _______________________________________________
>>>> Cytometry mailing list
>>>> Cytometry at lists.purdue.edu
>>>> https://lists.purdue.edu/mailman/listinfo/cytometry
>>>>
>>>>
>>>
>>>
>>>
>>> Simon Monard
>>> FACS Facility Manager
>>> MRC Centre for Regenerative Medicine
>>> Institute for Stem Cell Research
>>> University of Edinburgh
>>> Roger Land Building
>>> West Mains Road
>>> Edinburgh
>>> EH9 3JQ
>>>
>>> Tel. Lab 0131 6505876
>>> Tel Office 0131 6517265
>>>
>>> --
>>> The University of Edinburgh is a charitable body, registered in
>>> Scotland, with registration number SC005336.
>>>
>>>
>>> _______________________________________________
>>> Cytometry mailing list
>>> Cytometry at lists.purdue.edu
>>> https://lists.purdue.edu/mailman/listinfo/cytometry
>>
>> _______________________________________________
>> Cytometry mailing list
>> Cytometry at lists.purdue.edu
>> https://lists.purdue.edu/mailman/listinfo/cytometry
>>
>>
>> _______________________________________________
>> Cytometry mailing list
>> Cytometry at lists.purdue.edu
>> https://lists.purdue.edu/mailman/listinfo/cytometry
>> _______________________________________________
>> Cytometry mailing list
>> Cytometry at lists.purdue.edu
>> https://lists.purdue.edu/mailman/listinfo/cytometry
>>
>>
>
>
>
> Simon Monard
> FACS Facility Manager
> MRC Centre for Regenerative Medicine
> Institute for Stem Cell Research
> University of Edinburgh
> Roger Land Building
> West Mains Road
> Edinburgh
> EH9 3JQ
>
> Tel. Lab 0131 6505876
> Tel Office 0131 6517265
>
> -- 
> The University of Edinburgh is a charitable body, registered in
> Scotland, with registration number SC005336.
>
>
> _______________________________________________
> Cytometry mailing list
> Cytometry at lists.purdue.edu
> https://lists.purdue.edu/mailman/listinfo/cytometry



More information about the Cytometry mailing list