	Hi Betsy,

Watch out with venting a hood to outside: Our Lab Operations people 
tell me that for example in New York state this is illegal. So you'd 
need to check regulations first....


Kind regards,

Jan Hendrikx


>Betsy, we've had the Baker hood for a few years now, and its in a 
>very small room and is not ducted to building exhaust, and we have 
>trouble with heat buildup inside the hood.  If you are in a larger 
>room (this was reportedly the smallest room they've installed it in) 
>and can vent some of the heat away by connecting it to building 
>exhaust, you should be ok. Make sure to leave enough room for the 
>wet cart outside of the hood, too--its possible to put it under the 
>table inside the hood but definitely not the best setup to do it 
>that way.
>Mike Waring
>Ragon Institute of MGH, MIT, and Harvard
>
-- 




*********************************************************************************