John:

John wrote:
>     I am curious if the geometric mean is acceptable for calculating the 'area' of a bimodal or highly skewed (or pick an ugly distribution) population which is not easily discriminated? I acquire in log so I'm hesitant to use the arithmetic mean. Also, I use FCS Express and I'm trying to avoid exporting every gate in listmode and doing it the hard way...
>
>   

Geometric and arithmetic means estimate central tendency of a 
distribution, not an "area". Arithmetic mean is a reasonable measure of 
central tendency for unimodal, symmetric distributions, such as Gaussian 
distribution. Geometric mean estimates central tendency of log-normal 
distribution. If you can demonstrate that your data can be approximated 
reasonably well by log-normal distribution than you can use geometric 
mean. Of course, if the intensity values are represented in logarithmic 
fashion that you would have to calculate the log-average by computing 
the arithmetic mean of the logarithm transformed values. Anyway, you 
have to remember that mean is not a robust estimator - it  is very 
sensitive to extreme values. Trimmed mean, trimean, or median are 
usually better choices.

If your data shows bimodal distribution it is likely that what you see 
is a combination of two underlying normal (or log-normal) distributions. 
In this case reporting any single estimator is meaningless.

You may consider an alternative way of summarizing your data - instead 
of reporting mean values (e.g. for treated and untreated sample), you 
could report the distance between your samples and positive/negative 
controls. You can use for instance modified chi-square (1), KS-distance 
(2), or quadratic-form distance (3) as your measure of samples' 
(dis)similarity.

(1) Roederer, Mario, Adam Treister, Wayne Moore, and Leonore A. 
Herzenberg. 2001. Probability binning comparison: A metric for 
quantitating univariate distribution differences. Cytometry 45, no. 1: 
37-46.
(2) Brescia, Francesca, and Maurizio Sarti. 2008. Modification to the 
Lampariello approach to evaluate reactive oxygen species production by 
flow cytometry. Cytometry Part A 73A, no. 2: 175-179.
(3) Bernas, Tytus, Elikplimi K. Asem, J. Paul Robinson, and Bartek 
Rajwa. 2008. Quadratic form: A robust metric for quantitative comparison 
of flow cytometric histograms. Cytometry Part A 73A, no. 8: 715-726.

Bartek Rajwa

-- 