Dear Jason,

I've never used PE-Cy5.5 but according to its spectrum, it has a  
significant crosstalk to the PE channel.
We use FITC, PE, PerCP, PE-Cy7 and APC 5-color combo and it works fine  
on our Partec CyFlow space. Depending on your filter setup in your  
FC500 you may consider PerCP instead of PE-Cy5.5.

Bests,

Pal Jakso