bandpass filter I to measure PercP

Pál Jáksó pal.jakso at aok.pte.hu
Thu Jan 31 03:06:51 EST 2008


Hi Jose,

As you can see here: http://www.bdbiosciences.com/spectra/, both  
filters collect approximately 63-64% of the PerCP emission at  
different parts of the PerCP spectra.
Your Aria and the LSRII configuration may be optimized for different  
fluorochrom combinations so they equipped with different filters for  
PerCP.

Bests,

Pal Jakso

University of Pecs
Faculty of Medicine
Dept. of Pathology
12. Szigeti str.
H-7643 Pecs, HUNGARY



On 2008.01.29., at 17:38, Jose Diaz Romero wrote:

>
> Dear Flowers,
>
> I am using the 695/40 bandpass filter supplied with the LSRII to  
> measure PercP in this machine. Recently our facility has acquired a  
> FACSaria that according to the User’s Guide should be equipped with  
> a 675/20 bandpass filter. Is this bandpass filter better than the  
> 695/40 to collect PercP emission? Could I use the 675/20 from the  
> Aria in the LSRII? And the more important question, Where is  
> supposed to be this filter? According to the User’s Guide the filter  
> should be provided in the accessory kit, but I checked with the  
> people in our FACS facility and they do not rembember any extra  
> filter supplied with the Aria.
>
>
> Thanks
>
>
> Jose
>
>
> Jose Diaz-Romero
>
> Institute of Pathology
>
> University of Bern
>
> Switzerland
>