We have a 5 laser SORP Aria that has UV, 488, 532nm and can get
acceptable DNA profiles for PI stained samples on all three lasers. It
has taken some time and considerable help from BD and their engineers to
obtain this. The blue laser which is at time position 0 has always
looked fine but UV and green +/- the extremes for time delay showed no
identifiable DNA profile when aligned by BD with their standard
alignment beads. We, collectively, found that the laser steering mirrors
for UV and green occasionally need a "tweak" to obtain good CV's but it
is usually not too far out, also that a large window extension, say 5us
or more, is needed although I realise this may have some impact if
trying to sort at high speed. I find that we also need to tweak the
steering mirrors on a known well stained DNA sample since beads do not
seem to align the machine optimally for DNA, although if aligned with
beads the BD DNA QC kit (calf thymocyte nuclei) looks fine? Current
thoughts are that large particles (ie whole cells) behave differently in
the flow cell than smaller particles beads or nuclei??? We get doublet
discrimination similar to our LSRII and VantageDiVa when the machine is
correctly set. 

Best wishes


Ian Titley PhD
Section of Haemato-oncology
Institute of Cancer Research
15 Cotswold Road
Surrey	SM2 5NG

Tel +44 (0)20 8722 4255

>>> Jodene K Moore <drjmoore at temple.edu> 06/02/2008 14:50 >>>
Hi Flowers-I am under the impression that due to the unique beam
geometry of the Aria that cell cycle analysis is not possible
on this instrument (due mainly to the reduced spot size that
makes doublet discrimination impossible). Has anyone been able
to successfully run cell cycle on the Aria? Thanks for the
input. Any suggestions appreciated.

Always plan ahead. It wasn't raining when Noah built the ark.

Go to http://calendar.yahoo.com/facslabfels to check the Flow Cytometry
Facility online
appointment calendar.
Jodi Moore, PhD
Flow Cytometry Facility Manager
Temple University, School of Medicine
Fels Institute of Cancer Research and Molecular Biology
Room 315, AHP
3307 North Broad Street
Philadelphia, PA 19140

Core Lab: 215-707-4119, Room 220
Office: 215-707-7709, Room 315
Sorter: 215-707-4043, Room 221A

Temple Flow Core Facility Website:

The Institute of Cancer Research: Royal Cancer Hospital, a charitable Company Limited by
Guarantee, Registered in England under Company No. 534147 with its Registered Office at
123 Old Brompton Road, London SW7 3RP.

This e-mail message is confidential and for use by the addressee only.	If the message is
received by anyone other than the addressee, please return the message to the sender by
replying to it and then delete the message from your computer and network.

More information about the Cytometry mailing list