Hi Steve,

So if you are limited to a violet trigon on your LSRII does this mean you can only get
three Qdot fluor's or can you excite some with your blue laser and use the blue laser
octagon for some qdot emissions simultaneously? Has someone configured an optimal
arrangement given the standard three Pmt trigon for violet and an octagon for the blue
laser?

Gene Pizzo 

-----Original Message-----
From: Steve Perfetto [mailto:perfetto at mail.nih.gov] 
Sent: Thursday, June 14, 2007 9:46 PM
To: cyto-inbox
Subject: Re: Choosing filter configurations for Qdots in Flow

Joanna,

Actually, the best strategy for each Qdot detector is the use of a sharp
dichroic mirror with a wide band pass filter. After much trial and error the
attached filter configuration illustrates this point. You can also visit our
web site at; 
http://www3.niaid.nih.gov/labs/aboutlabs/VRC/flowCytometryCoreLaboratory/
for additional information. If your current system has a green laser, the
QD525 will be very dim because of the narrow 515/20 filter. I would
recommend using QD545 with the filters as listed in the attachment.

SP


Stephen P. Perfetto, MS.,MT. (ASCP)
Manager, Core Flow Cytometry Facility
Vaccine Research Center, NIH
Building 40
40 Convent Dr., Room 5507
Bethesda, MD 20892-3015

email: sperfetto at nih.gov
Phone: (301) 594-8659

Disclaimer:

The information in this e-mail and any of its attachments is confidential
and may contain sensitive information.  It should not be used by anyone who
is not the original intended recipient.  If you have received this e-mail in
error please inform the sender and delete it from your mailbox or any other
storage devices.  The National Institute of Allergy and Infectious Diseases
(NIAID)  shall not accept liability for any statement made that are the
sender¹s own and not expressly made on behalf of the NIAID by one of its
representatives.




On 6/14/07 11:11 AM, "Roberts Joanna" <joanna.roberts at epfl.ch> wrote:

>  
> Dear Flow Cytometry Community,
>  
> I have a question about filter configurations for Qdots in flow cytometry=




More information about the Cytometry mailing list