Dear Enid,

To get around the variation from day to day, we run Spherotech's Rainbow
beads and generate a linear calibration curve. We apply this to all our
samples.  In addition, we also run our calibrite beads through at the
settings we're using for our sample run and then track both whether the
calibration curve is linear and whether it evens out the calibrite's raw
values. I've attached a file showing typical calibrite readings <<calibrite
starts 7.22.05.xls>> . This seems to work pretty well for us. By the way.
the rainbow beads are a lot steadier than the calibrite. We have a FACscan
by the way. Hope this helps. Judith

> ----------
> From: 	Enid Keyser
> Sent: 	Friday, November 11, 2005 2:30 PM
> To:	Cytometry Mailing List
> Subject:	Day to day variability.
> 
> Question:  We are running Calibrite FITC beads on a daily basis just to
> see
> the day to day variability on our FACSCalibur.	The FL1 voltage was
> set on
> the first bead run and has not been changed since.  The median values for
> fluorescence intensity are the following:
> Day 1:	865.96
> Day 2:	1000
> Day 3:	964.66
> We need to eliminate or reduce, as much as possible, this day to day
> variability.	Any suggestions?
> 
> Enid Keyser 
> 
> 