PE and PI together: Summary

Nancy nogden at globelaboratories.com
Mon Nov 7 19:11:53 EST 2005


Thanks to everybody for their help with the question of whether running PE and PI in the same sample is advisable.  There is some conflict of opinion, probably due to differences in application, but here is a summary of suggestions, all of which settled some of my misunderstandings:

  1.. Use 7-AAD instead (or TOPRO-3 with a red laser, or Sytox Green)
  2.. Use low concentration of PI instead (0.5 ug/mL) 
  3.. Gate out PI positive population, and then compensation isn't required at all (this may not work if PE stains very intensely)
  4.. It is incorrect to determine PI and PE in separate samples, because gating for viable cells then isn't directly possible, and so non-specific PE-binding to dead cells isn't taken into account.

Nancy Ogden, M.Sc
Research Associate
Chemokine Therapeutics Corp.
2386 East Mall, Suite 303
Vancouver, B.C., Canada
V6T 1Z3


Tel:  604-822-8848

Fax: 604-822-0212

 

 

------------------------------------------------------------------------

The information contained in this e-mail message may be privileged, confidential and protected from disclosure. If you are not the intended recipient, any dissemination, distribution or copying is strictly prohibited. If you think that you have received this e-mail message in error, please e-mail the sender at "nogden at globelaboratories.com"