Sorting - LS threshold?

Pál Jáksó pal.jakso at
Sun Jun 20 03:59:24 EST 2004

Hello Ray,

Yes it does because  increasing the treshold for FSC means that the 
cytometer won't see the debris but it does not mean that it's not there 
so you may sort the debris into your sample without any control.


Pal Jakso