If you look at the tech specs (can be donwloaded from the BD website) 
the max analysis rate for 8 par is 70K/sec;slower if more parameters.

Sort rate (70 psi, 90 kHz drop drive) at 25K/sec, 98% purity, 80% of 
poisson theoretical yield.

I think that the sort rate has been confused with the analysis rate 
(perhaps even by BD).

You can sort faster than the spec, but the yield will drop.


Marty Bigos
Director, Flow Core
Gladstone Institute of Virology and Immunology
Building 3 SFGH Rm 509

>Dear Michele,
>my rule of thumb for sort rates are usually the frequency divided by 
>4, sometimes 5 depending on cell type. I've never heard BD 
>professing sort rates of 70K/sec .  I sort lymphocytes on FACSDiva 
>at 15K/sec (65KHz)  and on the Aria at 20K ( I haven't gone higher 
>because the cell concentration provided hasn't been dense enough). 
>  Viabilty is good across the board.  Hope that helps
>Hello All,
>Does anyone know the maximum sort rate on the Aria to recover viable 
>cells ( Lymphocytes for example)? I know BD say 70000 events/ sec 
>but this is not for viable cells.
>I can sort up to 10000cells/ sec on a Vantage, is the Aria 
>significantly higher?
>Thanks in advance.
>Michele Black
>Director, Cell Analysis Facility
>Dept. of Immunology, Box 357650
>University of Washington
>Seattle, Wa 98195
>Phone: (206)685-3014
>  Fax: (206)543-1013
>  mblack2 at
>Gisele Knowles
>Cytometry and Scanning Microscopy Core Facility
>Sunnybrook and Women's Research Institute
>B wing 210
>Sunnybrook Health Sciences Centre
>2075 Bayview Avenue, Toronto, ON
>M4N 3M5
>Ph: 416- 480-6100 x 7282, 7284
>Fax: 416-480-4375

-------------- next part --------------
HTML attachment scrubbed and removed

More information about the Cytometry mailing list