Isabell,
I have always avoided systemic ethanol, since I feel this can cause accumulation of fixed
debris.  That's probably the clouding that Partec is referring to.  I think it's best to
sterilize with bleach (5-10%).  You can also use a mild detergent as a pre-treatment.
Always follow with sterile sheath solution.
Note . . . never run bleach or detergent through the in-line filter.  Remove the filter
and connect a bypass before cleaning.
A further note on the in-line filter . . . I've removed these from my Vantage and Elite
sorters, and sort sterile frequently.  I've had no problems with contamination, since all
sheath fluid is 0.22u filtered before use.  Also, I routinely clean the sorters
systemically . . . a clean machine runs better, and is easier to make sterile when
necessary.
MAK.

Isabell Schmitt wrote:

> Dear Fellow Flowers,
> I have a question regarding decontamination of tubes and cuvette in a FACS Analyser
> (Galaxy) equipped with a sorting cuvette: Partec does not recommend a 70 % ethanol
> solution for decontamination of the system prior to sorting as alcohol would lead to
> clouding of the cuvette after some time. On the other hand, other companies like BD
> recommend the use of a 70 % ehtanol solution for this purpose. What we now did was use
> a 2 % Hypochloride solution, but we are not sure weather this is good for the cuvette.
> If anyone has experience in getting the system sterile and keep the cuvette clear in
> the long run we would be very happy, if you could give us some tips!
>
> Thanks a lot in advance
>
> Isabel Schmitt
>
> Isabel Schmitt
> Paul Ehrlich Institut
> Abtl. Medizinische Biotechnologie
> Paul Ehrlich Straße 51 - 59
> D-63225 Langen
> Germany
> Telefon: +49 (0) 6103-775206
> Fax:       +49 (0) 6103-771255

--
Mark A. KuKuruga, Managing Director
University of Michigan Core Flow Cytometry
<http://www.cancer.med.umich.edu/flow_cytometry>
phone: 734-647-3216  fax: 734-936-7376
kukuru at umich.edu





More information about the Cytometry mailing list