We are not experiencing the background problems with either CB, marrow, or 
peripheral blood evaluation of CD34.  Our backgrounds are usually 0-0.05% 
(10,000 total  or 50,000 live gate acquisition) with FITC, PE, or PerCP 
conjugated flourochromes.  Our assay volume is 150-200uL containing titrated 
range 10^5 - 10^6 cells and titrated CD34, diluent is PBS+1%(v/v)goat serum 
+0.1%(w/v)sodium azide, incubation 30min /wet ice/dark.  We set up two negative 
controls : autofluorescence and protein concentration matched isotype control.  
We are using an amonium chloride lyse buffer, 3-4mL/15min/room temp/dark, 
sediment cells 800g/5min and wash with 3-4mL of the PBS solution.  A source of 
the high non-specific binding may be related to the type of tube that your are 
using; we use minisorp tubes from Nunc Inter-Med Inc (Denmark) as they are a 
polyethylene specially treated  to minimize protein adsorption onto the plastic 
and subsequent leaching during cell washing.


April G. Durett
Technology Transfer - Flow Cytometry
Blood & Marrow Transplantation
Univ Texas MD Anderson Cancer Center
adurett at notes.mdacc.tmc.edu@internet

To: cyto-inbox
cc:  
From: yagrawal @ messi.uku.fi (Yashpal Agrawal) @ INTERNET
Date: 11/21/96 03:52:12 PM
Subject: staining of cord blood CD34+ cells


Dear flowusers,
  Problem:I have been having problems with CD34 cell
staining ofwhole Cord Blood(CB). Specifically the Isotype control (IgG1-PE from
 B.D) gives non-specific staining (0.2-0.6 % of mononuclear cells). This
small amount of non-specific staining causes difficulty in interpretation
when the CB stained with the specific HPCA-2-PE (also from BD) gives a
positivity in  0.5 % of cells and subtracting the background gives 0%
positivity. 

Do others have the same problem, any suggestions/advice is welcome ?.

Methods used:We collect our CB in blood bags (CPDA) and the sample is
stored at 4 C until  analyzed which is 1-4 hrs. Staining is according to
standard methods. 50 ul of CB + 20 ul antibody is incubated for 15 min in
cold and in dark. After this 2 ml FACSlysis solution (BD) is added,
incubated for 10 min and centrifuged for 5 min at 400 g. The cells are
then washed twice in PBS and analyzed immediately. 


Thanking you all,
Yash Pal Agrawal
Kuopio, Finland
yagrawal at messi.uku.fi

 