problem with harvesting cells after sorting

cawort01@roentgen.bcc.louisville.edu cawort01 at roentgen.bcc.louisville.edu
Fri Mar 15 10:55:15 EST 1996



what are you sorting the cells into?  if you are using plastic tubes you are getting what I jokingly call "like-Liked".  the tube will gradually pick up the charge of the droplets.  and when that charge gets high enough your sorted cells will be repelled by the tube.(not a good situation<G>).  all you need to do is use glass tubes.   I hope this helps.


Chris